Cleanliness Testing with ionograph

Views: 3277

My OMEGA Meter 600 is no nlonger operational. The new unit t... - Mar 31, 2009 by jhowrylak  

jhowrylak

#58464

Cleanliness Testing with ionograph | 31 March, 2009

My OMEGA Meter 600 is no nlonger operational. The new unit that I was looking at is the Ionograph 500. Does anyone have input on the Ionograph, or has anyone ever worked with both the Ionograph and the OMEGA600.

reply »

d0min0

#58466

Cleanliness Testing with ionograph | 31 March, 2009

Hi, we are using Ionograph 500 M since some time, what kind of info You need?

reply »

davef

#58477

Cleanliness Testing with ionograph | 31 March, 2009

Have you talked to people [Specialty Coating Systems 800-356-8260 317-244-1200 scscoatings.com/parylene_equipment/omegameter.aspx] about repairing your Omegameter?

Major differences between the testers are: * Ionograph measures conductivity, while the Omegameter measures resistivity. * Ionograph is a dynamic tester, while the Omegameter is a static tester. �Static� based systems actually are more susceptible to carbon dioxide adsorption than �Dynamic� based systems.